The di_maximum_rate
function sets the maximum Burst Rate for the device.
double di_maximum_rate(double max_burst_rate); |
The default maximum burst rate is the maximum burst rate allowed for the device.
Returns the maximum burst rate for the device that is closest to the requested value within the capability of the device. For some devices this is independent of the parameter.
This function should be called to avoid exceeding bandwidth capabilities of DI-720, DI-722, DI-730, and DI-78x products. The default setting for all DI-720, DI-722, DI-730, and DI-78x instruments is 250,000 (check your hardware documentation for sample rate specifications). For example, the maximum sample rate for a DI-722 is 50,000. To avoid exceeding bandwidth capabilities, max_burst_rate
should be set to 50,000 or lower.
Note: The Minimum value for Maximum Sample Rate should not be confused with the minimum value for di_sample_rate
, which is typically smaller by a factor of 32767.
Instrument | Maximum Sample Rate (default) | Recommended Maximum Sample Rate | Minimum Value for Maximum Sample Rate |
DI-145, DI-154, DI-195B, DI-194 | Baud Rate/201 | Baud Rate/201 | Baud Rate/201 |
DI-148, DI-158 | 14400 | 14400 | 450 |
DI-1492, DI-155 | 10000 | 10000 | 11.44427 |
DI-245 | 2000 | 2000 | 64.5161 |
DI-220, DI-221TC, DI-222 (Rev A or B) |
50000 | 50000 | 500 |
DI-222(Rev C or later) | 80000 | 80000 | 732.45 |
DI-1100 | 1600006 | 200006 | 10000 |
USB DI-1110, DI-1120, DI-2108, DI-4108, DI-4208, DI-4718B, DI-4730 | 1600003 | 200003 | 100003 |
Ethernet DI-4108, DI-4208, DI-4718B, DI-4730 | 1000003 | 125003 | 62503 |
DI-2008 | 20004 | 20004 | 64.51614 |
DI-2108-P | 1600004 | 400004 | 200004 |
DI-400, DI-401 | 500000 | 250000 | 500 |
DI-410 | 500000 | 100000 | 500 |
DI-700 | 976.5625 or 488.281255 | 976.5625 or 488.281255 | 976.5625 or 488.281255 |
DI-71x | 4800 | 4800 | 112.5 |
DI-500, DI-510, DI-5001, DI-720, DI-722, DI-730, DI-740 Printer Port Devices in Standard Mode |
200000 | 40000 | 500 |
DI-500, DI-510, DI-5001, DI-720, DI-730 Printer Port Devices in Bi-directional Mode |
200000 | 80000 | 500 |
DI-500, DI-510, DI-5001, DI-720 Printer Port Devices in EPP Mode |
200000 | 200000 | 500 |
DI-730 Printer Port Devices in EPP Mode |
200000 | 150000 | 500 |
DI-500, DI-510, DI-5001, DI-720 USB Devices |
DI-500, DI-510, DI-5001, DI-720 Ethernet Devices |
200000 | 180000 | 500 |
DI-722, DI-740 (all devices EXCEPT printer port devices in Standard mode) |
200000 | 50000 | 500 |
DI-730 USB and Ethernet Devices |
1 See Baud Rates.
2 DI-149 limits apply to the total sample rate, but arguments and return values for di_maximum_rate and di_sample_rate
are per channel values. With arguments of 1000 to both functions, all eleven channels may be enabled for a total throughput of 11000.
3 These values are maximum per channel and when multiplied by the numbers of channel enabled, the result should not exceed 160000. Call di_inlist
, di_maximum_rate
, and di_sample_rate
in that order.
4 Call di_inlist
, di_maximum_rate
in that order. For synchronized DI-2008 or DI-2108-P devices, this sets the maximum throughput per device.
5 488.28125 under Windows XP or Vista.
6 These values are maximum per channel, and when multiplied by (number of channels enabled plus 2) the result should not exceed 120000.
di_maximum_rate
can also be used to achieve sample rates lower than the maximum sample rate for the device divided by 32767 and to reduce cross-talk into channels driven by high-impedance sources. Although the Maximum Sample Rate for the device is the default when using DataqSdk.ocx, the SDK DLL user must call di_burst_rate
or rely on di_sample_rate
to do it indirectly before starting a scan.
If di_open
has not been called successfully or if the device has been closed, the return value for di_maximum_rate
is 0.0.
DI-1100, DI-1110, DI-1120, DI-2108, DI-4108, DI-4208 arguments and return values for sample rate are per channel values. Both di_list_length
and di_inlist
must be called first. Call di_maximum_rate
first, and di_inlist
again after di_sample_rate
to set the counter values.
error_code = di_open(1); |
/* for a device on parallel port LPT1 */ |
actual_max_rate = di_maximum_rate(50000.0) |
/* for a DI-722 */ |