Model DI-710-EHS
DI-710-EHS Stand-alone Ethernet Data Logger and Data Acquisition (DAQ) System, 14-bit, 14.4 kHz
- Includes a DI-710-EHS instrument, a 5-foot crossover Ethernet cable, a power supply, an NIST-traceable calibration certificate, a screwdriver for signal connections, and software on CD
Bundle and Save- 16 single-ended, 8 differential analog inputs
- Measurement range of ±1.25V to ±10V
- Allows in-the-field measurements to be saved to Secure Digital (SD) cards, or stored to a PC in real time using WinDaq software
- Built-in Real Time Clock allows automatic time and date stamping of data acquired to SD memory when used in the stand-alone mode.
- 14-bit analog to digital conversion.
- Includes WinDaq Software.
- Record up to 14,400 Hz in stand-alone mode using high speed SD card (model 101014-2G)
- Up to 4,800 Hz Sample Throughput Rate when connected to PC
- Unlock code required for sample rates over 1,108 Hz only when using WinDaq software
- Use HMP60 Relative Humidity (RH) and Temperature probe for environmental measurements
- Find out how long your data logger can record to SD memory using our Record Time Calculator
Description
DI-710 General-purpose Ethernet Data Acquisition System with Windaq software. Features 8 differential or 16 single-ended analog input channels, 4,000 Samples per second throughput, 14-bit measurement resolution, and full scale ranges of ±1.25, ±2.5, ±5, ±10 V full scale.
The DI-710 Series of products is a family of instruments for general purpose and stand-alone data logger data acquisition applications. Options include interface type, input voltage range, and PC-connected or standalone data logger operation. Interface options are USB or Ethernet. Gain ranges have selectable factors per channel of 1, 2, 4, and 8, or 1, 10, 100, and 1000. Instruments with the stand-alone data logger option feature a built-in multimedia socket that accepts standard Secure Digital (SD) memories to which acquired data may be stored without a connected PC. SD memories are the same commonly available mass storage devices used with digital cameras and MP3 players. Memories ranging in size from 16 MB to 2 GB are supported. Instruments without this option must remain tethered to a PC’s USB or Ethernet port during data acquisition and use the PC’s own program and memory to store acquired data.
All DI-710 Series products feature 14-bit measurement resolution, sixteen analog input channels that may be configured for single ended or differential operation per channel, and an 8-bit digital bi-directional port. PC-connected instruments stream data to the PC from as low as 0.048 Hz up to as high as 4,800 Hz throughput rate. Stand-alone data loggers store to their SD memory from as low as 0.0017 Hz up to as high as 14,400 samples per second. All DI-710 instruments are supplied with two removable, 16-position screw terminal access connectors.
Accessories

Description
USB Secure Digital Memory Card Reader connects to the USB port of any PC and allows data recorded to SD memory to be transferred to the PC for review and analysis.

Description
Unlocks WinDaq to allow recording speeds up to 4.8 kHz throughput. Unlock Code ONLY. No CD will be shipped. Please Note: If you are ordering this software for a device you already own, you MUST provide the serial number of your device under "Order Notes" during checkout to receive the unlock code. Please allow one (1) business day for processing.
Description
250 Ω ±0.1% Precision, Metal Film, 1/2W, Axial Resistor
Exceptionally low noise; typically 0.10 μV/V
Low voltage coefficient to ±5 ppm/V
Flame retardant coating. RoHS
compliant.

Description
Replacement 16-port removable screw terminal block for BOTTOM of screw terminal block receptacle in DI-78B, DI-710, DI-718B, DI-4718B, DI-718Bx, DI-785, DI-788, and DI-808 products.

Description
Replacement 16-port removable screw terminal block for TOP of screw terminal block receptacle in DI-78B, DI-710, DI-718B, DI-4718B, DI-718Bx, DI-785, DI-788, and DI-808 products.

Description
Power Connector for DI-1000TC, DI-710, and DI-718B products.

Description
Spare Australian Adapter for AC power supply 101017-2 and 101085.

Description
Spare UK Adapter for AC power supply 101017-2 and 101085.

Description
Spare European Adapter for AC power supply 101017-2 and 101085.

Description
Spare US Adapter for AC power supply 101017-2 and 101085.

Description
Pair of angle mounting brackets (with mounting screws) for DI-4718 Series, DI-4108 Series, DI-4208 Series, DI-808 Series, DI-710 Series, DI-718B Series, and DI-100TC Series instruments

Description
Re-calibration Service, this calibration certificate is included with new instruments. NIST-traceable certificate for all DI-710, DI-718B, and DI-718Bx data acquisition and data logger products. Includes both before and after data. Shipping charges do apply when ordering online. Learn more about Calibrating your device.
Software
WinDaq
Data Acquisition Recording and Playback Software
Included FREE with your DI-710 purchase. Installs drivers and WinDaq recording and playback software. Includes optional Add-on software with a valid installation key. 1108Hz maximum sample rate. For higher rates purchase the unlock code under the Accessories tab.

ActiveX Controls
Programming
FREE ActiveX programming solutions for Dataq Instruments products. Free sample programs available. Compatible with LabView.

WinDaq/XL
WinDaq-to-Excel Real Time Link
Real-time data logging directly to Microsoft Excel (supports MS Office 2003 or higher). No programming required. Free Trial Version. Purchase full version on WinDaq/XL product page.

Advanced CODAS
$595
Advanced analysis software add-on to WinDaq Waveform Browser software. Allows calculated channels, derivative, integration, moving average, automatic file management, automatic scaling, and more.

Specifications
| Analog Inputs | ||||||||||||||||||||||||||||||||||||||||||||||
| Number of Channels: | 16SE/8DI (software selectable per channel) | |||||||||||||||||||||||||||||||||||||||||||||
| Measurement Range Full Scale: |
|
|||||||||||||||||||||||||||||||||||||||||||||
| Input Impedance, single-ended: | 1MΩ | |||||||||||||||||||||||||||||||||||||||||||||
| Input impedance, differential: | 1MΩ each input to common | |||||||||||||||||||||||||||||||||||||||||||||
| Input bias current: | 10µA for a 10V input, single channel | |||||||||||||||||||||||||||||||||||||||||||||
| Input offset voltage: | Auto-zero | |||||||||||||||||||||||||||||||||||||||||||||
| Input offset current: | 2nA (single channel) | |||||||||||||||||||||||||||||||||||||||||||||
| Max. normal mode voltage: | 30V DC or peak AC | |||||||||||||||||||||||||||||||||||||||||||||
| Max. common mode voltage: | 30V DC or peak AC | |||||||||||||||||||||||||||||||||||||||||||||
| Common mode rejection: | 80db, gain=1, 1KΩ unbalance | |||||||||||||||||||||||||||||||||||||||||||||
| Channel-to-channel crosstalk rejection: | -75db @ 100Ω unbalance | |||||||||||||||||||||||||||||||||||||||||||||
| Gain temperature coefficient: | 50 ppm/°C | |||||||||||||||||||||||||||||||||||||||||||||
| Offset temperature coefficient: | 0.25µV/°C | |||||||||||||||||||||||||||||||||||||||||||||
| Digital filtering: | Standard: Conditional over-sampling Stand-alone: None |
|||||||||||||||||||||||||||||||||||||||||||||
| A/D Characteristics | ||||||||||||||||||||||||||||||||||||||||||||||
| Type: | Successive approximation | |||||||||||||||||||||||||||||||||||||||||||||
| Resolution: | 14-bit | |||||||||||||||||||||||||||||||||||||||||||||
| Monotonicity: | ±2 LSB | |||||||||||||||||||||||||||||||||||||||||||||
| Conversion Time: | 69µs | |||||||||||||||||||||||||||||||||||||||||||||
| Scanning Characteristics | ||||||||||||||||||||||||||||||||||||||||||||||
| Max. throughput sample rate: | Standard: 4,800 Hz (requires Unlock Code) Stand-alone: 14,400 Hz When acquiring more than one channel at a gain of 100 max throughput is 7200 Hz When acquiring more than one channel at a gain of 1000 max throughput is 900 Hz Dependent on SD card used. Low speed SD cards can sample up to 2,000 Hz High speed cards can sample up to 14,400 Hz. Some high speed cards cannot sample as high at 14,400 Hz but their capability can only be determined by trial and error (Model 101014-2G has been tested and approved). |
|||||||||||||||||||||||||||||||||||||||||||||
| Min. throughput sample rate: | Standard: 0.0034 Hz Stand-alone: 0.0017 Hz |
|||||||||||||||||||||||||||||||||||||||||||||
| Max. scan list size: | 17 entries | |||||||||||||||||||||||||||||||||||||||||||||
| Sample buffer size: | 2kb | |||||||||||||||||||||||||||||||||||||||||||||
| Controls (Stand-alone models) | ||||||||||||||||||||||||||||||||||||||||||||||
| Single push-button: | Manual control Record and Standby | |||||||||||||||||||||||||||||||||||||||||||||
| Calibration | ||||||||||||||||||||||||||||||||||||||||||||||
| Calibration cycle: | One year | |||||||||||||||||||||||||||||||||||||||||||||
| Digital I/O | ||||||||||||||||||||||||||||||||||||||||||||||
| Bits: | 8 bidirectional bits | |||||||||||||||||||||||||||||||||||||||||||||
| Configuration: | Each bit is programmable as Input or Output | |||||||||||||||||||||||||||||||||||||||||||||
| Output voltage levels: | Min. "1" 3V @ 2.5mA sourcing Max. "0" 0.4V @ 2.5mA sinking |
|||||||||||||||||||||||||||||||||||||||||||||
| Output current: | Max. source, -2.5 mA Max. sink, 2.5mA |
|||||||||||||||||||||||||||||||||||||||||||||
| Input voltage levels: | Min. required "1" 2V Max allowed "0" 0.8V |
|||||||||||||||||||||||||||||||||||||||||||||
| Ethernet Interface | ||||||||||||||||||||||||||||||||||||||||||||||
| Type: | 10/100Base-T | |||||||||||||||||||||||||||||||||||||||||||||
| Connector: | RJ-45 | |||||||||||||||||||||||||||||||||||||||||||||
| Protocol: | TCP/IP | |||||||||||||||||||||||||||||||||||||||||||||
| Server Type: | DHCP | |||||||||||||||||||||||||||||||||||||||||||||
| Removable Memory (Stand-alone models) | ||||||||||||||||||||||||||||||||||||||||||||||
| Type: | Standard SD (not SDHD or SDxC). Minimum speed of 13x. For sample rates above 2kHz SD speed should be 133x or higher. | |||||||||||||||||||||||||||||||||||||||||||||
| Capacity: | 16 MB to 2 GB | |||||||||||||||||||||||||||||||||||||||||||||
| Real Time Clock (Stand-alone models) | ||||||||||||||||||||||||||||||||||||||||||||||
| Type: | Date, hour, minute, second | |||||||||||||||||||||||||||||||||||||||||||||
| Resolution: | 1 second | |||||||||||||||||||||||||||||||||||||||||||||
| Accuracy: | 20 ppm | |||||||||||||||||||||||||||||||||||||||||||||
| Indicators | ||||||||||||||||||||||||||||||||||||||||||||||
| Stand-alone models: | Three-color LED indicating Record, Standby, and Error conditions | |||||||||||||||||||||||||||||||||||||||||||||
| Standard models: | Power LED | |||||||||||||||||||||||||||||||||||||||||||||
| Transfer Rate to PC | ||||||||||||||||||||||||||||||||||||||||||||||
| Real Time: | up to 4,800 samples per second | |||||||||||||||||||||||||||||||||||||||||||||
| From Memory Card: | up to 2,400 samples per second (Ethernet only) | |||||||||||||||||||||||||||||||||||||||||||||
| General | ||||||||||||||||||||||||||||||||||||||||||||||
| Panel indicators: | Mode LED | |||||||||||||||||||||||||||||||||||||||||||||
| Panel-accessible Power: | 5 VDC @ 10 mA max | |||||||||||||||||||||||||||||||||||||||||||||
| Panel Controls: | Control push button (Stand-alone models) | |||||||||||||||||||||||||||||||||||||||||||||
| Panel Slots: | Accepts SD-type flash memory | |||||||||||||||||||||||||||||||||||||||||||||
| Input connectors: | Two, removable sixteen position terminal blocks | |||||||||||||||||||||||||||||||||||||||||||||
| Operating Environment: | 0°C to 70°C | |||||||||||||||||||||||||||||||||||||||||||||
| Enclosure: | Aluminum base with steel wrap-around. Aluminum end-panels with plastic bezels. | |||||||||||||||||||||||||||||||||||||||||||||
| Dimensions: | 5.4375D × 4.125W × 1.5H inches (13.81D × 10.48W × 3.81H cm.) | |||||||||||||||||||||||||||||||||||||||||||||
| Weight: | 14 oz. | |||||||||||||||||||||||||||||||||||||||||||||
| Power Requirements: | USB (stand-alone): 9 to 36 VDC, 2 watts max USB (non stand-alone): 5V @ < 2 watts max (supplied by USB port) Ethernet: 9 to 36 VDC, 2.5 watts max |
|||||||||||||||||||||||||||||||||||||||||||||
Articles
Three Easy Steps for Stand-alone Data Logging
Stumbling Blocks to Stand-alone Data Logging
USB or Ethernet: Which Do I Choose?
Frequency and True RMS Data Acquisition
Intelligent Oversampling Methods Extend Measurement Flexibility
4-20mA Current Loop Data Acquisition
Powering Your Data Logger with a Battery
Ethernet Installation to DHCP network
Ethernet Installation to Static IP network
Ethernet Installation Directly-connected (one DI-71x to one PC)
IP Address Configuration for DI-71X-E Data Loggers
Six Ways To Deploy Model DI-71X Ethernet Data Loggers and Data Acquisition Systems
How to Manage Dynamic IP Addresses In Data Acquisition Applications Over the Internet
Videos
Introducing the DI-710
The DI-710 features 16 analog inputs and a variety of options to choose from.
DI-71XS Configuration
Learn how to set up and configure channels on a DI-71X for stand-alone data logging.
DI-71X Internet Installation
Learn how to connect to your DI-71X Ethernet over the Internet through your router.
DI-71X-E Direct Connection
Learn how to connect and install a DI-71XE directly to your PC with a Static IP address.


View Cart
sales@dataq.com
330-668-1444
Share